Current sensing of a low voltage surgical lighting system

Status
Not open for further replies.

dgreeley6

Member
Location
United States
We are trying to sense current (proxy for occupancy) of a Stryker low voltage surgical lighting system. The lighting control panel has no status output, totally encased cable from panel to light, and an incoming line voltage feed. Any ideas on HOW we could determine the light is "on" by CT or other method?
Thanks in advance.
 

GoldDigger

Moderator
Staff member
Location
Placerville, CA, USA
Occupation
Retired PV System Designer
Similar circuit to what is used in a power strip that senses current in the computer receptacle and cuts off other receptacles for accessories.
Not quite sure how they do that though.
You would need some sort of active circuitry, since you could not pull much power from a CT.

Tapatalk!
 

Jraef

Moderator, OTD
Staff member
Location
San Francisco Bay Area, CA, USA
Occupation
Electrical Engineer

steve66

Senior Member
Location
Illinois
Occupation
Engineer
I'm not sure how you would use one of those to sense current on a totally enclosed cable without breaking the cable open and cutting one lead to pass it through the sensor.

I wouldn't want to do that to an OR light.
 

GoldDigger

Moderator
Staff member
Location
Placerville, CA, USA
Occupation
Retired PV System Designer
We are trying to sense current (proxy for occupancy) of a Stryker low voltage surgical lighting system. The lighting control panel has no status output, totally encased cable from panel to light, and an incoming line voltage feed. Any ideas on HOW we could determine the light is "on" by CT or other method?
Thanks in advance.
What sort of indication do you need? Relay contact? Visual indicator?
Controlling other lighting or letting the security guard know the room is in use?


Tapatalk!
 

Jraef

Moderator, OTD
Staff member
Location
San Francisco Bay Area, CA, USA
Occupation
Electrical Engineer
I'm not sure how you would use one of those to sense current on a totally enclosed cable without breaking the cable open and cutting one lead to pass it through the sensor.

I wouldn't want to do that to an OR light.
Both companies offer a split core version that you open up, place around the conductor, and re-close. You would have to place it in an area where the cable, if sheathed, comes out of the sheath to terminate onto something.
View attachment 10067
 

Attachments

  • nk_technologies_atpr_series__large.jpg
    nk_technologies_atpr_series__large.jpg
    16 KB · Views: 0
Last edited:

GoldDigger

Moderator
Staff member
Location
Placerville, CA, USA
Occupation
Retired PV System Designer
I'm not sure how you would use one of those to sense current on a totally enclosed cable without breaking the cable open and cutting one lead to pass it through the sensor.

I wouldn't want to do that to an OR light.

You said that the cable from the control panel to the light is sealed, but that just means that the sensor would have to be placed on the AC input to the panel.
It would have to have a threshold between the light-on and light-off input current values to the panel.

Tapatalk!
 

steve66

Senior Member
Location
Illinois
Occupation
Engineer
You said that the cable from the control panel to the light is sealed, but that just means that the sensor would have to be placed on the AC input to the panel.
It would have to have a threshold between the light-on and light-off input current values to the panel.

Tapatalk!
Yes, that would work, and its much better than messing with the low voltage cable.

I guess that is what everyone else was thinking anyway, but for some reason, the title of the post made me think he was trying to put something on the low voltage side.
 